That Anne Frye stat irks me...
Colorflow and spectral doppler.. ie, visual doppler/ doppler ultrasound.. not handheld dopplers/ doptones are much more powerful then plain ultrasound.
Colorflow and spectral doppler are high frequency pulsed doppler using a transducer that is anywhere from 3-12Mhz, 3d and 4D US can be up to 2000x more powerful then a doppler...handheld doppler or doptone is a continuous wave at 2 or 3 mhz using a 9 volt battery
Yes, there is ultrasound, but no, the doppler your midwife uses is NOT more powerful then an ultrasound.. much less 20 or 30+ x more powerful.
http://www.naturalchildbirth.org/nat...ventions27.htm
[Vol 2 Issue 9 March 3, 2000]:
Do most midwives regularly make use of the Doptone? Also, shouldn't each
woman be made fully aware that the Doptone is ultrasonic? Even though the
exposure periods are usually brief, I heard that 1 min. of Doptone is much
stronger--equal to 30 min of the other full-image type ultrasound. Is this
true?
Anon.
====
[Vol 2 Issue 10 March 10, 2000]:
...
Second, you stated that the frequency of Doppler devices is higher than an
imager device. This is also not true. The Hitachi unit ... has a
trans-abdominal probe that transmits at 3.5 MHz. The industry standard for
obstetrical Dopplers is either 2 or 3 MHz. Imex in fact sells both. Higher
probe frequencies that we sell such as 5 and 8 MHz are used for vascular
applications and are labeled "Not Designed For Fetal Use."
In addition, I would like to point out that the FDA limit for power
intensity emitted by a continuous wave ultrasound for fetal use is 0.094
watts per square centimeter. The FDA power intensity limit for pulsed wave
ultrasound for fetal use is 190 watts per square centimeter. The power
emitted by a Doppler can be 2,000 times less than an imager! Imex 3 MHz
probes emit 0.009 watts per square centimeter, a factor of ten times less
than the FDA limit.
Reprinted from Midwifery Today E-News (Vol 2 Issue 10 March 10, 2000)